The two best UV/X-ray sensors that have fast time response (<1 ns) and are radiation hard are GaAs sensors sold by Nutrek and diamond sensors sold by Diamond detectors limited in the UK and Alameda Applied Sciences Corp in the US. The calibration of each detector varies from unit to unit but both are extremely radiation hard. Nominal values for diamond sensitivity are 1E-4 A/W above the band gap of ~5 eV.
Another thought, if you are using Si, you are saturating your detector. The total charge read out is preserved but the time history is incorrect. If you are working on a yield sensor, your Si might be fine. If you are interested in time response, you need to limit the input aperture or switch to a less sensitive material.